aluminum pipe for helium quench

aluminum pipe for helium quench

aluminum pipe for helium quench

(OP)
  I am working on the quench pipe for a liquid helium cooled magnet.  (We need 300mm inside diameter and can only have 0.4 bar of back pressure for ~126ft length of pipe.) Right now I am thinking aluminum.  Have been searching for a good design reference - ie: which alloys are suitable, connections suitable for these conditions, system contraction, etc.   

RE: aluminum pipe for helium quench

Hi dmi,
The correct code for this is ASME B31.3 in the US.  Don't know what it is in other countries.  If US, aluminum alloys are the best material and are rated to the -452 F temperature you need.  Stress allowables start on page 182 (ASME B31.3-2006).  They also have values for thermal contraction.  Designing for thermal contraction is important.

Another issue you should be aware of is that uninsulated pipe will drip liquid air, so there should be nothing flamable beneath the pipe.  It will also drip water, so having electrical equipment beneath the pipe is a no-no.  You can insulate the pipe, and it should last a long time since quenches are relatively rare.  You might also put drip pans beneath the pipe.  Either way, keeping liquid air and water from dripping on things is desireable.  

Note that the back pressure MRI manufacturers provide is probably not something you can calculate with any accuracy.  They may provide a peak flow rate (flow starts at zero, spikes when the burst disk releases, drops, then increases again before dropping exponentially).  Heat transfer during this event is too complicated to calculate with any accuracy, so I'd suggest putting the pipe in they suggest and not worry about doing a pressure drop calculation.  If you do want to do the dP calculation, you can be conservative and use a relatively high temperature (say -100 F).  

Another concern is where it is venting.  Flow can be very high, and the helium is both extremely cold and an asphixient.  Plan on the vent location accordingly.

Why 300 mm though?  Most installations I've seen are 4" to 8" pipe.  Don't think I've ever seen one at 12".  

RE: aluminum pipe for helium quench

(OP)
Thanks for the reference and yes I'm in the US.

The pressure(.4bar)is from the manufacturer's rep and he did the calcs to sized the pipe once we had a route.  I imagine it's a little larger because we need several elbows before we can get outside.

Are these types of pipes generally welded?  

RE: aluminum pipe for helium quench

I don't know what's the most common, but I suspect it's welded.  My experience with these is one step away from the actual installation.  I deal primarily with the manufacturing environment (MRI mfg) and recovery of gas at the plants.  

Just out of curiosity, what is the max flow rate you are designing to?  Can you say what magnet and manufacturer?  Just curious....

RE: aluminum pipe for helium quench

(OP)

The total volume for the unit is 355 liters.  Not positive on max flow rate during a quench.  

It's a Bruker FTMS.

RE: aluminum pipe for helium quench

I have seen these in Al, Cu, austenitic stainless steel, Ni alloys (expensive), titanium (including some alloys), and Invar.
Always welded.
Managing stress from expansion/contraction is the real issue.

Don't overlook Ti.  It will have fairly low CTE (only Invar will be lower) and this diameter could be fabricated.

= = = = = = = = = = = = = = = = = = = =
Plymouth Tube

RE: aluminum pipe for helium quench

(OP)
I'd never heard of Invar.  Looks interesting.

Thanks.

 

Red Flag This Post

Please let us know here why this post is inappropriate. Reasons such as off-topic, duplicates, flames, illegal, vulgar, or students posting their homework.

Red Flag Submitted

Thank you for helping keep Eng-Tips Forums free from inappropriate posts.
The Eng-Tips staff will check this out and take appropriate action.

Reply To This Thread

Posting in the Eng-Tips forums is a member-only feature.

Click Here to join Eng-Tips and talk with other members!


Resources