Is there a rule of thumb to follow when choosing a calibrator?
Is there a specific accuracy or other specification one should strive for?
For instance, if calibrating a DMM how much more accurate must the calibration meter be?
Windo
RE: Is there a rule of thumb to follow when choosing a calibrator?
The traditional MIL-STD-52B benchmark was 10:1 test accuracy ratio (TAR), but more recently, the trend has been a 4:1 TAR.  This makes a reasonable compromise, since independent calibrator errors would therefore would only contribute about 3% to device's measurement error.
The closer you get to require device accuracies comparable to NIST calibration limits, the lower the TAR that you can achieve.
TTFN
TUR (Test Uncertainty Ratio) in modern terminology.
If you use a TUR of 10:1 you tend to ignore the uncertainty in the calibrator completely. At 4:1 you may then want to consider "guard banding". there are some excellent articles on the web on this subject. Google "TUR" and "guard band".