BTU Output of Centrifuges
BTU Output of Centrifuges
(OP)
Does anyone know how I can measure or caculate the BTU output of a few centrifuges that we have in our labs. Apparently, if the output is above 10% of the other equipment I will ahve to do a room mapping.





RE: BTU Output of Centrifuges
RE: BTU Output of Centrifuges
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
"It's the questions that drive us"
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
RE: BTU Output of Centrifuges
RE: BTU Output of Centrifuges
RE: BTU Output of Centrifuges
RE: BTU Output of Centrifuges
If the samples are of large quantity AND immediately removed however, then you would have to somehow determine how much heat transferred into them, maybe just a simple little differential measurement.
JRaef.com
"Engineers like to solve problems. If there are no problems handily available, they will create their own problems." Scott Adams
For the best use of Eng-Tips, please click here -> FAQ731-376
RE: BTU Output of Centrifuges
If you just have an on-off switch though, and a mechanical brake, then all the above is true, every bit of power you send to your motor will be converted to heat in the room with a 100% efficacy.
RE: BTU Output of Centrifuges
RE: BTU Output of Centrifuges
RE: BTU Output of Centrifuges
RE: BTU Output of Centrifuges
Do these units have a nameplate that sez: 120V @ 2A or 300W or ?
Keith Cress
Flamin Systems, Inc.- http://www.flaminsystems.com