Establishing a ground connection on a portable resistance test unit

Establishing a ground connection on a portable resistance test unit

Establishing a ground connection on a portable resistance test unit

(OP)
I recently purchased a Megger DLRO200 low resistance ohmmeter. Unfortunately the unit must be plugged into a 120 VAC source which is not convenient in our manufacturing facility. We purchased a pure sine wave inverter (300 W), battery charger and 12 V battery in hopes of converting to a self contained unit we could roll around on a cart. The problem we are facing is that the DRLO performs a test to verify a ground connection. The output on the inverter has a ground prong but aparrently not adequate for the DRLO to pass the test. I'm looking for any suggestions on how to get the DRLO into thinking it has a ground connection.

Thanks in advance,

RE: Establishing a ground connection on a portable resistance test unit

Are you looking for a 'portable ground' ?!?

RE: Establishing a ground connection on a portable resistance test unit

Low resistances? neutral

----------------------------------
  Sometimes I only open my mouth to swap feet...

RE: Establishing a ground connection on a portable resistance test unit

(OP)
After getting to the right person at Megger I found out that I would have to establish a ground through the auxiliary ground terminals of the unit. A bit on an inconvenience but it will work.The ground connection is simply a precaution to assure that the circuit being tested is safe to test.

Thanks to all for their posts

Red Flag This Post

Please let us know here why this post is inappropriate. Reasons such as off-topic, duplicates, flames, illegal, vulgar, or students posting their homework.

Red Flag Submitted

Thank you for helping keep Eng-Tips Forums free from inappropriate posts.
The Eng-Tips staff will check this out and take appropriate action.

Reply To This Thread

Posting in the Eng-Tips forums is a member-only feature.

Click Here to join Eng-Tips and talk with other members!


Resources