Testing of large capacitors
Testing of large capacitors
(OP)
There are a number of large capacitors in power supply circuits on our trains. Typical sizes include 3300uF @ 350V; 1uF @ 660V, 10uF @ 800V.
Does anybody have any recommendations for test equipment to check if capacitors are good, or starting to fail.
A capacitance check on a multimeter will measure the capacitance using millivolts, and I am not sure that this is an accurate representation of what happens when these devices have several hundred volts across them.
Are there any other checks that should be made to test capacitors? (DC Resistance?, Equivalent Series Resistance?) How are power factor correction capacitors checked?
Does anybody have any recommendations for test equipment to check if capacitors are good, or starting to fail.
A capacitance check on a multimeter will measure the capacitance using millivolts, and I am not sure that this is an accurate representation of what happens when these devices have several hundred volts across them.
Are there any other checks that should be made to test capacitors? (DC Resistance?, Equivalent Series Resistance?) How are power factor correction capacitors checked?





RE: Testing of large capacitors
1. Visit
http://www.gallawa.com/microtech/cap_test.html
http://www.evinfo.com/list.html
http://www.pv.infn.it/~servel/atlas/hv/hedgehog/
http://www.novacap.com/NOVACAPTECHNICALBROCHURE.pdf
etc. for more info
2. Also, one may follow industry standards, e.g. IEEE, IEC, etc.
http://www.ieee.org
IEEE Std 18-1992 IEEE Standard for Shunt Power Capacitors
http://www.iec.ch
RE: Testing of large capacitors
RE: Testing of large capacitors
after several minutes at the operating voltage.
<nbucska@pcperipherals.com>
RE: Testing of large capacitors
RE: Testing of large capacitors
I have had great sucess using a capacitor analyzer made by Sencore and no I don't work for the company. It saved my bacon many times over when trouble shooting power supplies, and circuits.
I have seen so many techs/engineers use a multimeter to test caps and conclude that they were good and then could not find anything else wrong with the device under test. It is a very common mistake.
For high power caps I would suggest possibly Hi pot/Doble/megger. I have not tried any of these myself but I'm sure if you contact the manufac. they will be able to point you in the right direction.
Good luck.
RE: Testing of large capacitors
RE: Testing of large capacitors
The Sencore LC103 looks like the instrument for the job.
For those interested in the method used to date, we wired a 40W, 110V lamp in series with the capacitor under test and a variac and isolation transformer/rectifier, and slowly increased the voltage across the capacitor up to working voltage. We would then remove the supply voltage, and measure the time it took for the voltage to drop 50 volts. Anything less than 1 minute was a reject. If the capacitor had an internal flashover, the lamp would limit the current flowing and protect the power supply.
I am sure the proper instrument will provide more accurate results.
Rodney.
RE: Testing of large capacitors
1.insulation test,use meg meter.
2.capacity test,apply voltage(from 0 to full rate),read the current and caculate its capacity.
!when apply voltage be sure current in a limited value.
!after test,don't forget discharge the cap.
torontoele
RE: Testing of large capacitors
Here's how it worked. btw, this was an analog tester.
Short the 2 leads of the tester together. This measurement is the same measurement you should get when testing a capacitor (Don't rely on this method using caps smaller than 10uF). It should appear as if the cap is shorted (even though it's not...check to make sure the cap is not shorted either using an ohm meter). If the value is significantly different, then the cap is bad.
RE: Testing of large capacitors
<nbucska@pcperipherals.com>